Hi Mike,
We had the standard xenon-arc lamp (#6251NS, Newport), but we've recently changed to the ozone-free xenon-arc lamp (#6263, Newport). The reason we changed to the ozone-free lamp is occupant health, as it reduces ozone generation from UV below 260nm. Above 260nm, the irradiance of the #6251NS and #6263 lamps appears similar, so we apply a UV filter and a hot mirror to block UV.
Regarding power, we use 75W lamps. We operate our Whitmore system at around ~800 mlm, and 75W was enough to reach the light level. Also, we feel like we don't need to expose our samples to a higher light dose. However, when a light bulb deteriorates, we increase our wattage slightly higher with a digital controller.
Our Whitmore system also has a water filter; however, we don't use it due to the complication of connecting to the DI water. We spoke to Newport, and they suggested removing it if we don't use it, since it can reduce the light output. We use a hot mirror to cut the IR, and it seems sufficient.
I hope my answers are helpful, and I am happy to discuss more if you have any follow-up questions.
------------------------------
Jenny Kim
Assistant Scientist
Getty Conservation Institute
Los Angeles
United States
------------------------------